i have used vectrino velocitmeter for a few weeks now. but by last week, the velocimeter keeps give me error status
please have a look of the attachment image (in the words file), how can i fix it?
am i using it for too long? because it was very warm in the lab during summer time in australia
and also, my probe doesn't sense the temperature as well
how am i be able to fix it?
many thanks
ray
Hi Ray,
I see from the screenshot that you use a very old firmware. You should upgrade to the latest firmware 1.29. You can download it from http://www.nortek-as.com/en/support/software. You should also download the latest Vectrino PC software. Please do this first and report the results!
Best regards
Marius
Thank you Marius for your help
and also i want to ask
sometimes i've got negative velocity readings
i tried to measure the velocity for a period of time, thought the negative sign may disappear. however, the negative sign is still there no matter how long i wait. but if i change the height of the probe to another different height, the velocity readings are more logical
what could be the problem?
please see attachment
Ray
Ray,
In your attached plot your correlation values are extremely low. In general the correlation values should be upwards of 90%. Please check your Nominal Velocity Range setting in the Deployment Planning dialog and ensure it is high enough for the flow you are measuring. In this same dialog on the right hand side, two velocity ranges are reported, one each for horizontal and vertical velocities. The Nominal Velocity Range setting will cause changes in the reported horizontal and vertical ranges. Make sure these are above the expected flow velocities or you will not measure usable, high quality data. I would suggest moving up to a higher Nominal Velocity Range to start.
P.J.
Thank you very much !
also, because of this velocimetre is a few years old, and my supervisor couldn't locate the sofeware licence key.
am i be able to question the supplier for the licence key?
if anyone can help me on this
ray
Thanks Peter
i have already emailed them the serial number and the probe number
so do you think the error i am having (see the very first attachment) is related to the firmware version?
see if i upgrade the firmware, the error status will not appear?
ray