Burst mode

Up to Velocimeters - single point

Burst mode

Posted by mehrdad at September 16. 2009

Hi,

Would you please explain me that how the burst mode work in Vector? More specifically, my question is as follows:

When I fix the Burst Interval to 6 and sampling per burst to 10, does that mean that the instrument will collect 10 data at a second with 6 second interval? Then, if it would be true, what is the effect of frequency at this case?

Regards,

Mehrdad

 

Re: Burst mode

Posted by P.J. Rusello at September 16. 2009

Hi Mehrdad,

In burst mode you specify two parameters, Burst Interval which is the time between when bursts start and Samples Per Burst, which sets the number of samples collected in the burst.

In the example you gave, there would be a new burst every 6 seconds which would contain 10 samples. The Rate parameter still controls how fast the samples are taken. If you set the rate to 64 Hz, the 10 samples would take 0.15 seconds to be collected. If you set Rate to 1 Hz, you would receive an error since 10 samples can not be collected in 6 seconds at a rate of 1 Hz.

 

P.J.

 

Re: Burst mode

Posted by mehrdad at September 18. 2009
Thanks for the explanation.

Re: Burst mode

Posted by mehrdad at September 18. 2009
Dear Peter, I also want to make sure that I understood a part of advance deployment planning correctly. If I fix the Sampling Volume to 22 and Transmit length to 8, it means that the height of sampling volume is 22 mm. So if the distance between sampling volume and bottom boundary (which is being reported by Vector) would be 50 mm, it means that the bottom of sampling volume is [50-(22/2)] mm away from the boundary since the Vector reports the distance to the center of sampling volume. Is that right? Knowing this calculation is important for me the time which I want to take data near boundary. Regards, Mehrdad

Re: Burst mode

Posted by P.J. Rusello at September 21. 2009

Yes, the Vector will report the distance to the center of the sample volume so the volume will extend half of the height above and below this value. The Vector should report the distance to the boundary from both the sample volume and central transducer (or probe as labelled in the data files) at the beginning and end of the measurement cycle.

 

Re: Burst mode

Posted by mehrdad at September 28. 2009
Thanks again peter